    HP 54520A does it support XY mode?

    HP 54520A. Does it support XY mode? Please help

  2. #2
    Super Moderator
    Points: 251,556, Level: 100
    Awards:
    1st Helpful Member

    Join Date
    Jan 2008
    Location
    Bochum, Germany
    Posts
    43,804
    Helped
    13320 / 13320
    Points
    251,556
    Level
    100

    Re: HP 54520A does it support XY mode?

    Looks like it doesn't.


    1 members found this post helpful.

    •   AltAdvertisment

        
       

  3. #3
    Advanced Member level 5
    Points: 15,909, Level: 30
    Achievements:
    7 years registered
    neazoi's Avatar
    Join Date
    Jan 2008
    Location
    Greece
    Posts
    3,226
    Helped
    12 / 12
    Points
    15,909
    Level
    30

    Re: HP 54520A does it support XY mode?

    Quote Originally Posted by FvM View Post
    Looks like it doesn't.
    Is that quite common throughout DSOs that they do not have XY mode, like analogue scopes do?



  4. #4
    Advanced Member level 4
    Points: 8,141, Level: 21
    Achievements:
    7 years registered

    Join Date
    Nov 2009
    Location
    Italy
    Posts
    1,111
    Helped
    383 / 383
    Points
    8,141
    Level
    21

    Re: HP 54520A does it support XY mode?

    They usually have a math function to do this. In your case it could be the "versus" function.


    1 members found this post helpful.

    •   AltAdvertisment

        
       

  5. #5
    Super Moderator
    Points: 251,556, Level: 100
    Awards:
    1st Helpful Member

    Join Date
    Jan 2008
    Location
    Bochum, Germany
    Posts
    43,804
    Helped
    13320 / 13320
    Points
    251,556
    Level
    100

    Re: HP 54520A does it support XY mode?

    In your case it could be the "versus" function
    Right, unlike modern DSOs that have X-Y mode selectable under horizontal settings, 54520A has it under math functions.


    1 members found this post helpful.

  6. #6
    Advanced Member level 5
    Points: 15,909, Level: 30
    Achievements:
    7 years registered
    neazoi's Avatar
    Join Date
    Jan 2008
    Location
    Greece
    Posts
    3,226
    Helped
    12 / 12
    Points
    15,909
    Level
    30

    Re: HP 54520A does it support XY mode?

    Quote Originally Posted by FvM View Post
    Right, unlike modern DSOs that have X-Y mode selectable under horizontal settings, 54520A has it under math functions.
    Yes, that is right, it is the VER (versus) function. In the same menu, it also has FFT function, it can be converted into a spectrum analyzer up to it's limit of 500MHz. Resolution is about 1.3KHz.
    I have tested the FFT and although quite slow at it's full resolution (because of the FFT calculation) it is useable. I am so excited!
    I hope it is quite reliable in dbm measurements. If I remember well, by testing the same signal on the scope function the Vpp is not exactly as it should be expected (it is greater) based on this table. http://www.neazoi.gr/conv.pdf Could it be a calculation error or miscalibration?

    An analogue scope that I owned gave more Vpp amplitude on the same signal generator I tested, but I was testing it on 1Mohm. Right now I am testing it on 50 Ohm on the 54520A, giving a lower amplitude signal, so I hope it is more accurate.
    Last edited by neazoi; 25th November 2013 at 09:44.



    •   AltAdvertisment

        
       

  7. #7
    Advanced Member level 4
    Points: 8,141, Level: 21
    Achievements:
    7 years registered

    Join Date
    Nov 2009
    Location
    Italy
    Posts
    1,111
    Helped
    383 / 383
    Points
    8,141
    Level
    21

    Re: HP 54520A does it support XY mode?

    Your table is correct and the conversion between volt and dBm is valid only for a 50 ohm terminated system. The measurement with the scope in high impedance you should have the voltage doubled with respect to the scope terminated on the same impedance of that of the generator. If your generator is 50 ohm impedance and you see 1 Vpp with the scope terminated onto 1 Mohm then you will see 50 ohm/(1Mohm+50 ohm)*1Vpp≈0.5 Vpp.

    Could you post the values you have read: Vpp on the scope and dBm from FFT ?


    1 members found this post helpful.

  8. #8
    Advanced Member level 5
    Points: 15,909, Level: 30
    Achievements:
    7 years registered
    neazoi's Avatar
    Join Date
    Jan 2008
    Location
    Greece
    Posts
    3,226
    Helped
    12 / 12
    Points
    15,909
    Level
    30

    Re: HP 54520A does it support XY mode?

    Quote Originally Posted by albbg View Post
    Your table is correct and the conversion between volt and dBm is valid only for a 50 ohm terminated system. The measurement with the scope in high impedance you should have the voltage doubled with respect to the scope terminated on the same impedance of that of the generator. If your generator is 50 ohm impedance and you see 1 Vpp with the scope terminated onto 1 Mohm then you will see 50 ohm/(1Mohm+50 ohm)*1Vpp≈0.5 Vpp.

    Could you post the values you have read: Vpp on the scope and dBm from FFT ?
    The fft measures 4.5-5dbm and I should expect a maximum of 1.125vpp, but when switching to scope I get a little bit more at 1.4vpp.
    Is that normal?



  9. #9
    Advanced Member level 4
    Points: 8,141, Level: 21
    Achievements:
    7 years registered

    Join Date
    Nov 2009
    Location
    Italy
    Posts
    1,111
    Helped
    383 / 383
    Points
    8,141
    Level
    21

    Re: HP 54520A does it support XY mode?

    No, 3 dB difference seems to me too much. In case of miscalibration I would expect a difference between generator setting and scope measurement, but good agreement between trace and FFT.
    Could you say what about the frequency of your signal, the RBW of the FFT, the horizontal set of the scope (total acquisition time and sampling time) and the windowing type (flat top, Blackmann Harris, etc.)



  10. #10
    Advanced Member level 5
    Points: 15,909, Level: 30
    Achievements:
    7 years registered
    neazoi's Avatar
    Join Date
    Jan 2008
    Location
    Greece
    Posts
    3,226
    Helped
    12 / 12
    Points
    15,909
    Level
    30

    Re: HP 54520A does it support XY mode?

    Quote Originally Posted by albbg View Post
    No, 3 dB difference seems to me too much. In case of miscalibration I would expect a difference between generator setting and scope measurement, but good agreement between trace and FFT.
    Could you say what about the frequency of your signal, the RBW of the FFT, the horizontal set of the scope (total acquisition time and sampling time) and the windowing type (flat top, Blackmann Harris, etc.)
    Hello,
    Sorry it took me so long to respond.
    I did some tests and here are the results.

    1. In scope mode:
    1.61vpp

    2. In FFT mode:
    FFT1 (Hamming - frequency resolve) 7.5dbm
    FFT2 (Rectangular - transitions) 5.8dbm
    FFT3 (Flat top - amplitude acc.) 8.5dbm

    Any thoughts?



--[[ ]]--