HP 54520A. Does it support XY mode? Please help
Follow along with the video below to see how to install our site as a web app on your home screen.
Note: This feature may not be available in some browsers.
Looks like it doesn't.
In your case it could be the "versus" function
Right, unlike modern DSOs that have X-Y mode selectable under horizontal settings, 54520A has it under math functions.
Your table is correct and the conversion between volt and dBm is valid only for a 50 ohm terminated system. The measurement with the scope in high impedance you should have the voltage doubled with respect to the scope terminated on the same impedance of that of the generator. If your generator is 50 ohm impedance and you see 1 Vpp with the scope terminated onto 1 Mohm then you will see 50 ohm/(1Mohm+50 ohm)*1Vpp≈0.5 Vpp.
Could you post the values you have read: Vpp on the scope and dBm from FFT ?
No, 3 dB difference seems to me too much. In case of miscalibration I would expect a difference between generator setting and scope measurement, but good agreement between trace and FFT.
Could you say what about the frequency of your signal, the RBW of the FFT, the horizontal set of the scope (total acquisition time and sampling time) and the windowing type (flat top, Blackmann Harris, etc.)