Please i need a very good clarification in regards to the maximum input voltage of an osscilloscope.
i just bought one the Rigol1074z which came with the Rigol 2200 probe. the datasheet says at 10x maximum input voltage is 300ACV but i am not really cleared about this if it is RMS, Peak or peak to peak voltage.
so i tried mailing the Rigol support and here is the response I got
"The RP2200 probes that come with the DS1000Z series of scopes are rated as CATII probes with maximum voltage of 300VAC. This is 300V RMS. This does include any DC offset as well.
The maximum voltage is also frequency dependent. I have included the RP2200 specifications guide for your reference.
If you are going to be working with circuits that have amplitudes above 200V, it may make sense to use probes that have a higher voltage rating."
there are two things that are very unclear to me.
1. "This 300V RMS this does include any DC offset.
2. "If you are going to be working with circuits that have amplitudes above 200V, it may make sense to use probes that have a higher voltage rating. Does this means i cant use the probe on 10x to measure AC RMS voltage up to 300V
Here is the link to the datasheet of the probe
https://www.batronix.com/pdf/Rigol/RP2200.pdf
please help me out.
thanks
Tunde
Please i need a very good clarification in regards to the maximum input voltage of an osscilloscope.
i just bought one the Rigol1074z which came with the Rigol 2200 probe. the datasheet says at 10x maximum input voltage is 300ACV but i am not really cleared about this if it is RMS, Peak or peak to peak voltage.
so i tried mailing the Rigol support and here is the response I got
"The RP2200 probes that come with the DS1000Z series of scopes are rated as CATII probes with maximum voltage of 300VAC. This is 300V RMS. This does include any DC offset as well.
The maximum voltage is also frequency dependent. I have included the RP2200 specifications guide for your reference.
If you are going to be working with circuits that have amplitudes above 200V, it may make sense to use probes that have a higher voltage rating."
there are two things that are very unclear to me.
1. "This 300V RMS this does include any DC offset.
2. "If you are going to be working with circuits that have amplitudes above 200V, it may make sense to use probes that have a higher voltage rating. Does this means i cant use the probe on 10x to measure AC RMS voltage up to 300V
Here is the link to the datasheet of the probe
https://www.batronix.com/pdf/Rigol/RP2200.pdf
please help me out.
thanks
Tunde