Continue to Site

Welcome to EDAboard.com

Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

HP 33120 function generator output

Status
Not open for further replies.

winglj

Full Member level 4
Joined
Feb 21, 2005
Messages
195
Helped
2
Reputation
4
Reaction score
2
Trophy points
1,298
Activity points
3,063
hp33120

Hi, I have a question about the HP 33120 function generator.

When the output was set to 100mV p-p value in the generator, why the oscilloscope shows that the waveform is 200mV p-p value? Thanks in advance.
 

hp33120 specification

I believe that this problem would not have to happen, can be a problem of the resolution of the oscilloscope.
It can that depends on the load to which these applying the signal, if this is next to a short circuit for example.
You can see the specification in this pdf.
 

    winglj

    Points: 2
    Helpful Answer Positive Rating
Your HP 33120 is set to 50-ohm output impedance, and your oscilloscope is set to high-impedance.
You can change the HP 33120 mode:
MENU -> SYS_MENU -> OUT_TERM -> 50_OHM or HIGH_Z

That setting doesn't actually change the output impedance. It simply changes the scale factor.
 

    winglj

    Points: 2
    Helpful Answer Positive Rating
Status
Not open for further replies.

Part and Inventory Search

Welcome to EDABoard.com

Sponsor

Back
Top