Welcome to our site! EDAboard.com is an international Electronic Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Register Log in

How to measure differential signals using an Oscilloscope

Hesambook

Member level 3
Joined
Sep 26, 2007
Messages
59
Helped
1
Reputation
2
Reaction score
2
Trophy points
1,288
Activity points
2,091

Differential signal measurement in practice

To measure a differential signal, we have two options, one is using a differential probe and second is using a two channels oscilloscope. A differential probe is expensive but handles a better accuracy. Using two/four channels oscilloscope is the cheapest method which handles acceptable results.

Just you need to connect oscilloscope’s Channel-1 to one of the differential lines/wires and Channel-2 to another one. Then go to the math function and enable CH1-CH2 which means a difference. Then adjust the oscilloscope to observe the signal. You use the run/stop or a single shot button to freeze the signal on the screen and examine it.

I have examined the ADSL2+ signal on the telephone line using the above-mentioned differential measurement method.

Reference: http://bit.ly/2Z7sDTt
 

BigBoss

Advanced Member level 5
Joined
Nov 17, 2001
Messages
4,745
Helped
1,438
Reputation
2,874
Reaction score
1,301
Trophy points
1,393
Location
Turkey
Activity points
28,663
This method is a bit subjective and unstable.If the frequency is sufficiently high, phase shifts will be different by oscilloscope channels so the measurement may be a bit erroneous.
You presume that the oscilloscope has %100 identical channels and Phase Shifts are linear but this is not true in reality..
 

Hesambook

Member level 3
Joined
Sep 26, 2007
Messages
59
Helped
1
Reputation
2
Reaction score
2
Trophy points
1,288
Activity points
2,091
This method is a bit subjective and unstable.If the frequency is sufficiently high, phase shifts will be different by oscilloscope channels so the measurement may be a bit erroneous.
You presume that the oscilloscope has %100 identical channels and Phase Shifts are linear but this is not true in reality..
I am agree, as I mentioned in the article using the differential probe is the best, but not necessarily high voltage differential probes, because the main problem is to deal with high frequency pulses. This method at least gives a close picture of the final signal. The result on the screen is the result of the difference
 

schmitt trigger

Advanced Member level 5
Joined
Apr 17, 2013
Messages
2,646
Helped
844
Reputation
1,692
Reaction score
819
Trophy points
1,393
Activity points
17,756
Also....one has to make sure that both probes are properly and equally compensated.

A good way I have found is to compensate each probe individually (with the scope's CAL SIGNAL output).
Then attach both probes simultaneously to the same CAL output and perform the A-B comparison, and slightly tweak for maximum cancellation.
 

dick_freebird

Advanced Member level 5
Joined
Mar 4, 2008
Messages
6,760
Helped
1,991
Reputation
3,986
Reaction score
1,815
Trophy points
1,393
Location
USA
Activity points
54,208
Also....one has to make sure that both probes are properly and equally compensated.

A good way I have found is to compensate each probe individually (with the scope's CAL SIGNAL output).
Then attach both probes simultaneously to the same CAL output and perform the A-B comparison, and slightly tweak for maximum cancellation.
I like to use the "signal of interest" (or one side of the pair)
for this because the CAL signal on most 'scopes is way lower
frequency / edge rate than most of what I look at.

You would of course like to get identical model# probes to
start with.
 

pjmelect

Advanced Member level 2
Joined
May 13, 2012
Messages
521
Helped
120
Reputation
240
Reaction score
119
Trophy points
1,323
Activity points
4,259
On older oscilloscopes you invert one channel and sum the two channels. One problem with doing this is that the bandwidth is drastically reduced to around 1Mhz depending on the scope.
 