Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

How our body can induce noise?

Status
Not open for further replies.

fala

Full Member level 5
Joined
Sep 18, 2005
Messages
249
Helped
19
Reputation
38
Reaction score
4
Trophy points
1,298
Activity points
3,569
Hello, I've recently built an amplifier circuit with overall gain of 1000000. after shortening two differential inputs to ground and connecting output to an oscilloscope, I encountered an interesting phenomena. when I was stand still in a few meters of the amplifier, oscilloscope was showing a straight line(no noise) but when I was moving my hand in about a meter from amplifier, oscilloscope was very noisy. because the gain was 1Mega so noise voltage should be in matter of a few microvolts. can anybody explain it completely? If I'm acting as a capacitor as a friend earlier suggested how it can be explained that with a meter distance I can induce a few uVolts noise?
 

There are many possibilities based on the information you have given:

1. Static electricity charge/discharge. If the air in the room is relatively dry, waving your arm will change the electrostatic charge on your body. The minute static discharge arcs in your clothing may be producing broadband noise picked up by the loop formed from the scope leads to the circuit.

2. Changes in mangetic flux through the scope lead loop to and from the circuit under measurement. Your body is a conductor, and you are moving it through the earths magnetic field in the vicinity of the measurement probe loop. A moving conductor in a magnetic field generates minute curents which, in turn, cause changes in the surrounding field. It's a weak field, and the currents are small, but your amplifier has a large gain.

3. Your body is acting as a reflector for incident RF. There are many sources of RF in our homes - commercial AM and FM radio, TV, shortwave broadcast, computer CPU clocks, microwave ovens, radar, etc. Any non-linearity in the contacts for your scope measurement would act as a rectifier (demodulator). Waving your arm would change the incident RF fields in the region of the scope probe loop. You would see that as a change in the character of the 'noise' displayed on the scope.

Try characterizing the noise - is it broadband, or is it within a narrow range of frequencies? Is it random amplitude, or can you reproduce signals within a defined band of amplitudes? Are you just changing existing noise level, or are you actually creating noise when you wave your arm.

I could go on, but you need to do more measurements to narrow down the possibilities. Shielding your circuit and the scope connections would be a good start. You could then separate out the noise introduced directly into the circuit from the surrroundings, from the noise indroduced through power sources or other connections. Keep in mind that the noise can also be coming from your scope itself.
 