pisoiu
Advanced Member level 3
The probe should be compensated in order to obtain a flat response over its designed bandwidth. Otherwise, what you see as 1Vpp@13,56Mhz may not be accurate. Usually all scopes have an output exactly for that purpose. The output generate rectangular signal, and you should change the variable capacitor from the probe (with a plastic screw driver) untill you see a nice rectangle, without overshoots and undershoots. Usually, all these operations are explained in detail in probe's manual. Remember, the compensation is when the probe is set to 10:1, not at 1:1.
The difference between the readings at 1V and 10V may come from your generator, it may be out of calibration.
/pisoiu
The difference between the readings at 1V and 10V may come from your generator, it may be out of calibration.
/pisoiu