Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Need Help to select Data logger

Status
Not open for further replies.

Kunal barot

Newbie level 3
Joined
Jul 10, 2009
Messages
4
Helped
0
Reputation
0
Reaction score
0
Trophy points
1,281
Location
mumbai
Activity points
1,304
Dear Friend

I need to purchase a data logger. I have 4 thermocouple and 2 RTD input.
I am confused what should be the minimum sampling Rate require for this application. ( Or what should be the timimg interval)
 

I am confused what should be the minimum sampling Rate require for this application.
You need to define the sampling rate you need for your particular application. For glaciers, I sample once per hour. For Ramp-and-hold oven temperature tracking, I may sample every second.

Ken
 

Like KMoffett said it depends of your application.

I think the good question you have to ask to yourself is:
What is the quickest event I want to be able to detect with my logging ?
Then choose a shorter sampling rate (2 times shorter is generally good).

Hope that can help you
 

Thanks For replay Friends.

As thermocouple and PT 100 both response time is nearly 1 sec. so it is better to select data loggger with sampling speed of 1 sec or 500 msec

Like KMoffett said it depends of your application.


I think the good question you have to ask to yourself is:
What is the quickest event I want to be able to detect with my logging ?
Then choose a shorter sampling rate (2 times shorter is generally good).

Hope that can help you
 

You don't have too consider the response time of the probe but the response time of the all system.
Is the temperature you're going to log is susceptible of these variations ?
In brief if you intend to log a 1000 L water tank temperature you don't have to log every half second, as the temperature of the water will not vary that fast.
I don't know if you catch what I mean.
 

what i mean , let suppose My theromocouple response time is 1 Sec .
so thre is no meaning to select data logger sampling speed less than 1 sec because therocouple temerature reading will updated after 1 sec.
 