Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

what more accurate, power meters or VNA?

Status
Not open for further replies.

SloppyMagic

Newbie level 3
Newbie level 3
Joined
Nov 26, 2012
Messages
3
Helped
0
Reputation
0
Reaction score
0
Trophy points
1,281
Visit site
Activity points
1,343
Hello everyone,

I'm new here so I hope I'm posting this in an appropriate forum (well actually i have reposted it in this forum having only just discovered it, sorry for the duplicate post)

Here is the question

I need to know if measuring a 40 dB loss ( a ku band 40 db W/G coupler for example) would be more accurate using a suitable VNA (Agilent 8722 for example) or would making the measurement using two power meters be more accurate? Freq range is 13750 MHz- 14500 MHz

what I am interested in here is amplitude only, no phase requirement.

I contend that power meters are the most accurate method of measuring power. However, some of my engineering colleagues just love their VNAs.

In order to do this calibration measurement I would employ an Agilent 8481A sensor to measure an incident power of aprox 5-10 dBm and then use an Agilent 8481D to measure the DUT output, in this case the coupling arm of the 40 dB coupler. The 8481A is suitable for measuring levels from +20 to - 30 dBm, but you don't really want to operate in the top 10dB of any of the Agilent 848X series detector's dynamic range (where they suffer some linearity error).

Likewise, the output signal of the secondary arm of the coupler would be measured using a second power meter employing the hi sensitivity 8481D sensor (-20 to -70 dBm).

if we assume an incident signal level of +5 dBm on the 8481A this would produce a signal level of aprox -35 dBm on the 8481D sensor/power meter (both power meters are Agilent 437B). This signal level keeps both sensors out of the top 10 dB of their respective dynamic ranges for max accuracy.

The measurement is done in two steps. software divides the freq band into 401 points and steps the synthesized source (Agilent 83624A) through 401 CW freqs while making an incident measurement at each CW step.

The DUT (the coupler in this example) is then inserted and, with the 8481D sensor on the secondary arm, the 401 CW freq step is repeated while measuring the signal from the coupler secondary arm using the 8481D and a second 437B. After that a little bit of arithmetic will provide the coupling loss.

The measurement using the 8722 would just be a simple through calibration followed by insertion of the DUT coupler.

I understand the VNA would be much faster, but speed is not a consideration here, while accuracy is paramount. i also understand that any change in signal level between the incident and DUT measurement with the power meters constitutes direct error so GREAT care is exercised during the process to ensure a stable incident signal.

You can assume all measurements are made using correctly configured instruments with valid and traceable calibration. Coax to waveguide adapters are Maury Microwave low VSWR adapters (VSWR <1.05, insertion loss <0.1 dB) and the mainline of the coupler is terminated in a low vswr waveguide load (again, less than 1.05:1 over the freq band of interest)

So which is likely to be more accurate in this scenario, the twin power meter measurement or the VNA?

thanks for any inputs

regards

Ed.
 

Hello everyone,

I'm new here so I hope I'm posting this in an appropriate forum (well actually i have reposted it in this forum having only just discovered it, sorry for the duplicate post)

Here is the question

I need to know if measuring a 40 dB loss ( a ku band 40 db W/G coupler for example) would be more accurate using a suitable VNA (Agilent 8722 for example) or would making the measurement using two power meters be more accurate? Freq range is 13750 MHz- 14500 MHz

what I am interested in here is amplitude only, no phase requirement.

I contend that power meters are the most accurate method of measuring power. However, some of my engineering colleagues just love their VNAs.

In order to do this calibration measurement I would employ an Agilent 8481A sensor to measure an incident power of aprox 5-10 dBm and then use an Agilent 8481D to measure the DUT output, in this case the coupling arm of the 40 dB coupler. The 8481A is suitable for measuring levels from +20 to - 30 dBm, but you don't really want to operate in the top 10dB of any of the Agilent 848X series detector's dynamic range (where they suffer some linearity error).

Likewise, the output signal of the secondary arm of the coupler would be measured using a second power meter employing the hi sensitivity 8481D sensor (-20 to -70 dBm).

if we assume an incident signal level of +5 dBm on the 8481A this would produce a signal level of aprox -35 dBm on the 8481D sensor/power meter (both power meters are Agilent 437B). This signal level keeps both sensors out of the top 10 dB of their respective dynamic ranges for max accuracy.

The measurement is done in two steps. software divides the freq band into 401 points and steps the synthesized source (Agilent 83624A) through 401 CW freqs while making an incident measurement at each CW step.

The DUT (the coupler in this example) is then inserted and, with the 8481D sensor on the secondary arm, the 401 CW freq step is repeated while measuring the signal from the coupler secondary arm using the 8481D and a second 437B. After that a little bit of arithmetic will provide the coupling loss.

The measurement using the 8722 would just be a simple through calibration followed by insertion of the DUT coupler.

I understand the VNA would be much faster, but speed is not a consideration here, while accuracy is paramount. i also understand that any change in signal level between the incident and DUT measurement with the power meters constitutes direct error so GREAT care is exercised during the process to ensure a stable incident signal.

You can assume all measurements are made using correctly configured instruments with valid and traceable calibration. Coax to waveguide adapters are Maury Microwave low VSWR adapters (VSWR <1.05, insertion loss <0.1 dB) and the mainline of the coupler is terminated in a low vswr waveguide load (again, less than 1.05:1 over the freq band of interest)

So which is likely to be more accurate in this scenario, the twin power meter measurement or the VNA?

thanks for any inputs

regards

Ed.
In short, to measure Power, the Power Meter is better, but you can not see
power at specific frequencies. A VNA, or even a Spectrum Analyzer can better
give you a better understanding of Power at different Frequencies. Also the VNA
can give you impedance / S11, S22.

So if you want to know the Power as in a PA measurement Power meter is best
if you are looking to find out about matching/performance use a VNA.
 

Measuring attenuation by the combination of two power meters sounds like provoking extra measurement uncertainty. But for an exact answer you need to go into the specification details of the involved instruments. VNA is clearly the straightforward way, 8722 magnitude measurement uncertainty is specified with about 0.2 dB in the range of interest.

Both measurement setups can be calibrated by a precision attenuator, if you have one.
 
  • Like
Reactions: pewang

    pewang

    Points: 2
    Helpful Answer Positive Rating
I would go for power meter. Is more stable measurement, more reliable, more repeatable, and much cheaper than VNA.
 
  • Like
Reactions: pewang

    pewang

    Points: 2
    Helpful Answer Positive Rating