Voltage Measurement varying between Devices?

Status
Not open for further replies.

digi001

Full Member level 5
Joined
Apr 5, 2011
Messages
244
Helped
22
Reputation
44
Reaction score
21
Trophy points
1,298
Activity points
2,904
Hi I'm looking to measure Voltage RMS but having trouble finding which device in my lab is most accurate.

I measure the same 440V Line to Line signal with a:

Tektronix DPO - 440V (Samples at 100,000Hz)
Fluke - 446V (Not sure)
Dranetz Power Analysis - 446V (Samples around 6000Hz)

What should I trust?
 

To compare results, a bandwidth specification may be required. It's meaningless to count e.g. MHz interferences for Vrms.

Oscilloscope specifications have limited accuracy by nature, e.g. 2%, this also applies to derived measurements. Review your manual. In so far, you should trust the dedicated voltage instruments, particularly the power analyzer.
 
Reactions: digi001

    digi001

    Points: 2
    Helpful Answer Positive Rating