Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

VNA calibration-measurement

Status
Not open for further replies.

Julew

Member level 1
Joined
Apr 7, 2011
Messages
40
Helped
1
Reputation
2
Reaction score
1
Trophy points
1,288
Activity points
1,586
Hello!

To measure the S11 of a LNA, manufacturer says that the generator output power of VNA has to be -30dBm so I can´t configure the attenuator of VNA so i bought an external one (20dB). Today i try to measure S11 but i can´t calibrate the equipment with the attenuator (if i do this, when i connect load/oc/sc it goes crazy!) (i attach load pictures) and if i attach it after calibration S11 of the attenuator is taken into account and the measure is not valid. How can make this measurement?
 

Attachments

  • 50 ohm fase.png
    50 ohm fase.png
    26.8 KB · Views: 80
  • 50 Ohm.png
    50 Ohm.png
    126.1 KB · Views: 84
  • 50 ohm mag.png
    50 ohm mag.png
    25.6 KB · Views: 80

I already mentioned in reply to a previous post of your's, that a S11 measurement with an external attenuator doesn't work. Now you verified this statement in your test.

To allow a reduced level S11 measurement, a VNA must be equipped with an internal attenuator between source and reflected signal measurement. (or an option to connect an external attenuator to the internal signal path). When connecting the attenuator between VNA and DUT, you'll effectively measure the attenuator's rather than the DUT's S11.
 

Why ??
Because a 20dB attenuator after VNA will provide 40dB isolation and that's why Output Port of your VNA will see approx. 50 Ohm that is almost independent from your load or calibration standard.
In additional to this, -30dBm output power will increase your uncertainty..
 

Ok but, why i get this crazy smith chart? It has to be pointing to 50 Ohm ....
 

Ok but, why i get this crazy smith chart?
Most likely it's just noise. Possibly, the VNA calibration fails, because the software isn't prepared for effectively zero difference between open, short, load. If you use a more moderate attenuation, e.g. 6 dB, you should get at least plausible results.
 