op-amp problem
Sinisa , Thanks for your reply.
I will clarify the test setup further :
There are 3 electrodes as shown in my previous post.
Electrode A is platinum of 5mm sq.
Electrode B is a calomel electrode
(
https://en.wikipedia.org/wiki/Saturated_calomel_electrode)
Electrode C is a steel plate of 10mm sq.
The electrolyte is a common salt solution 3%V/V in a glass jar,wherein all the three are immersed. There are no other objects in the cell .
I connected R75 ( 10M,1% ) to ground. I used a 10M instead of 1M to reduce the loading on the calomel electrode.
With no power given to the cell ( i.e. no voltage to Electrode 1),
At pin 1(IC18A) I get 0.09V , At pin8 (IC18C) I get 0.094.
The LCD display on my unit varies from 0.103~0.086 .( If I short the inputs my display is steady at 0.004 ~ 0.010.)
With the power connected and voltage set to 1.5V , the circuit maintains 1.484 ~ 1.531 as measured by the ADC.This variation "seems" to be more due to noise than the effect of the control action.If I touch the multimeter probe the voltage drops to 0.9V before the control action again boosts the cell voltage to maintain the 1.5V.
The drive circuit for powering the cell is discussed here :