Monitoring HV
engr_joni_ee
Advanced Member level 3
How do we monitor high voltage at the level of 1000 to 1200 V that we use for Photo Multiplier Tube (PMT) which don't require much current. I guess in micro-amp range. Let's say 250 micro-amp.
The ADC usually accept input from 0 to 3.3 V. How do we monitor HV ?
The photomultiplier has a negative cathode potential, right? So, buffer your ADC input
with a big current-limit meter resistor, into an inverting op amp configuration. That
allows you to diode-clamp the op amp input pin to GND without disturbing the measurement.
Something like -.001 voltage gain, will get you directly to the ADC input. Rail-to-rail
op amp or rail-to-rail with attenuator resistors (if the op amp power supply exceeds
ADC tolerance) would be suitable.
It sounds bad to use an amplifier to attenuate, but the input impedance of the ADC is
possibly low, and you need it driven (but don't want to load the HV supply).
The signal conditioner design depends on your specified response curves and tube design.
See Chapt. 5 https://web.archive.org/web/2014050...u.com/resources/pdf/etd/PMT_handbook_v3aE.pdf
H2431 - 50 circuit diagram in Figure 5-13 look like a internal circuit of the PMT having dynode chain inside.
I would like to measure HV using voltage divider outside PMT. I was thinking if I chose
R1 = 2.7 Kohm
R2 = 1000 Kohm
Voltages will be across them
Vo = R1 / (R1 + R2) * 1000 = 2.69 V
Vo = R1 / (R1 + R2) * 1200 = 3.23 V
Current will run through them
I = V / R
I = 1000 / (1000 k + 2.7 k) = 0.99 mA
I = 1200 / (1000 k + 2.7 k) = 1.19 mA
How do I calculate power rating for resistors R1 and R2.
Power across R1 + R2 will be.
P = V I = 1000 x 0.99 mA = 990 mWatt
P = V I = 1200 x 1.19 mA = 1428 mWatt
The classic approach is to use a high-resistance voltage divider. Choose resistors with high voltage ratings and ensure the total resistance is at least ten times the PMT's internal resistance to minimize loading effects. For example, with a 1 MΩ PMT resistance, you could use two 10 MΩ resistors in series (20 MΩ total). The voltage across the lower resistor (connected to the ADC) will be 1/20th of the HV.
Hi lenzaca99,
You mentioned to chose the resistors with high voltage ratings. What voltage rating for the resistors do you suggest to measure the HV at the level 1000 - 1200 V ?
Regarding the total resistance to be at least ten times the PMT's internal resistance. You mean the resistors in the voltage divider should be at least ten times the PMT's internal resistance ?
Let's consider the internal resistance of PMT is 1 MΩ then how I can chose voltage divider resistors to get 3.3 V ?