Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.
Below 1nH I would go for an accuracy no better than +/-30% (maybe worst).
Inductance measurement in a PNA is an indirect measurement, calculated from an impedance measurement.
You can use a VNA to measure inductance. You just need to measure at a high enough frequency so 1 nH presents a impedance near 50 ohms. That also means you need a good calibration. You can use either 1 port or 2 port measurements, preferably both.
The 4287A is just a 1-port VNA anyway. The bridge type LCR meters are probably too low in frequency to measure 1 nH, however they are very accurate for certain ranges.
You don't have to measure the inductance directly. You can resonant it with a known capacitance and measure the suck out. There are methods using transmission lines too. One method may be better a finding the Q, another better for L or SRF. You need to research the different methods.
In this case, I would prefer to measure L and Q over wide frequency band. Hence using LC oscillator is not suitable for my application.
But madengr gives a very good idea: Is it if |Zin| near to 50ohm, measurement accuracy can be greatly improve, assume inductor are measured using 2ports VNA?