Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

[SOLVED] Energy Meter using ADE7758

Status
Not open for further replies.

Neveth

Newbie level 6
Joined
Aug 17, 2009
Messages
11
Helped
0
Reputation
0
Reaction score
0
Trophy points
1,281
Location
Coimbatore, India
Activity points
1,372
Guys, i'm doing a energy meter using ADE7758. I've a problem on calibrating it.
Anyone have done it successfully ? Help me !
I've done the communication(SPI) successfully. And also i'm getting some value from it. I dont' know whether it is right or wrong !
I need help to calibrate the rms readings, 7758 gives a unsigned 24 bit rms value, what i've to do with it ? What should i load in the VRMSGAIN and VRMSOS(offset) registers ? how can i cross verify the result ?

Thanks in advance.
 

you want to count the RMS value from the meter right?

---------- Post added at 13:25 ---------- Previous post was at 13:22 ----------

In your meter How many leds are there ?

on that board what they mention for that LED ?

---------- Post added at 13:27 ---------- Previous post was at 13:25 ----------

tell me after that i will tell back how to calibrate?
 

Calibration means comparing to a known standard and adjust, if necessary. If you don't have a known accurate standard, e.g. a good RMS multimeter, you can't calibrate and should better rely on precalculated gain factors.
i'm getting some value from it. I dont' know whether it is right or wrong
Calibration means, that you need to get clear about right or wrong first.
 