I am facing an issue with ADS1115 ADC which is 16 bit ADC. I am able to communicate mircocontroller and ADC. I am even able to get the ADC counts. Though counts are fluctuating I have used averaging method to get proper counts.
The PGA is +/- 2.048. I am using channel 0. I am getting below results,
mV ADC count
200 200
300 300
1000 1000
1024 1024
1058 1058
To convert ADC count into voltage I am dividing ADC count by 1000. My question is how to get fourth digit after decimal point.
Suppose mV=1058.5 I need converted value should be 0.10585.
As the ADC is 16 bit, can anyone please explain me how to get fourth digit after the decimal point.
Your average values can have higher precision than the ADC count.
If you add 10 fluctuating ADC conversions and divide the sum by 10 you have a higher precision value with 0.1 mV resolution.
Over how many samples are you averaging now?
Are you using a "sliding window" average?
if you are using 16-bit data.
the least voltage you can measure is
Sensitivity (least voltage you can measure) = Voltage range / 2^16;
Sen = 4.096/65535;
Sen = 62.5 uV / bit
means you can read 62.5 uV on change of 1 bit.
your result should be as
Let your reading of 16-bit ADC is 32090.
Voltage = (32090/65535) * 4.096
Voltage = 2.0056 Volts
Also note that you really need a *VERY* clean board design, power supply etc. to measure voltages to that level of accuracy. In my experience,you can reliably get 8, maybe up to 10 bits, from an ADC before the noise takes over.
Don't expect to measure more accurately than the circuit will allow.
Susan
Maybe true for a hobbyist's design.
Read the datasheet where the limit is. If the ADC datasheet tells you that you get noise free results up to 15 bits then be sure that one can achieve this on a PCB.
But for sure you need careful design ... and the ADC can't output less noise than there is at the analog input.
My recommendation: less is more. If you add circuitry, especially OPAMPs, you add noise, drift and distortion.
Only add them where they are unavoidable.