I had posted this before in a different thread which is not related to the issue at hand. My apologies.
Time and frequency measurements work very well using the cursors, however, volts measurements are far off, perhaps 150 times larger than they should be. For example, a 0.5V PP square wave reads back 46 V. Cursor movements and displayed voltage are very smooth, just off by a large factor.
I set up the unit to measure DC with a variable power supply. I traced the output of the DC filter that is used by the comparator in the B trace section and found that it tracks the power supply changes very well. I also was able to verify that the measurement processor is able to put out a DAC voltage exactly the same as the input DC and it appears that the comparator output is zero with the ZERO HYST signal high. I know that the entire DAC, ADC and comparator systems are working, else the displayed voltage would not track the cursor movements so well.
It almost is as if the measurement processor thinks it is putting out a much larger voltage than it actually is and this much larger voltage is also displayed on the screen, ie, some sort of software error or bad EPROM where the voltage codes are stored or some such. I thought at first that this is a readout error (ie, displayed value error), but if that were the case, it would not go to OVERV after a displayed level of several hundred volts.
Everything else on the scope works very well. This is a tough error since nothing is pointing to a hardware issue. Perhaps I have a wrong version EPROM? Can someone help me understand the workings of the system or point me where I could diagnose further.
Thanks in advance.
Time and frequency measurements work very well using the cursors, however, volts measurements are far off, perhaps 150 times larger than they should be. For example, a 0.5V PP square wave reads back 46 V. Cursor movements and displayed voltage are very smooth, just off by a large factor.
I set up the unit to measure DC with a variable power supply. I traced the output of the DC filter that is used by the comparator in the B trace section and found that it tracks the power supply changes very well. I also was able to verify that the measurement processor is able to put out a DAC voltage exactly the same as the input DC and it appears that the comparator output is zero with the ZERO HYST signal high. I know that the entire DAC, ADC and comparator systems are working, else the displayed voltage would not track the cursor movements so well.
It almost is as if the measurement processor thinks it is putting out a much larger voltage than it actually is and this much larger voltage is also displayed on the screen, ie, some sort of software error or bad EPROM where the voltage codes are stored or some such. I thought at first that this is a readout error (ie, displayed value error), but if that were the case, it would not go to OVERV after a displayed level of several hundred volts.
Everything else on the scope works very well. This is a tough error since nothing is pointing to a hardware issue. Perhaps I have a wrong version EPROM? Can someone help me understand the workings of the system or point me where I could diagnose further.
Thanks in advance.