Hi All
I make a set of coaxial cable TRL calibration kit. but I have a little confuse of defining the parameter of calibration kit.
I use Agilent E5071C network analyzer. The below example was from Agilent website. It tells me I need to define the offset loss of thru and line. But in my understanding, TRL theory should not need any known parameter of calibration kit, even the delay of line.
It is a little weird, If it is necessary, how do I define it? there are two different coaxial cable from NA to DUT, one is "expensive cable" which it was used to being reference plan of SOLT calibration. And the other one is TRL calibration kit cable. so I have two different cables if I need to measure offset loss. Do I measure cable1, cable2 or cable1+cable2?
"
Example of defining the TRL calibration kit
You need to enter the definition of the TRL calibration kit to perform TRL calibration. Follow these steps to define the following calibration kit given as an example.
THRU (Delay 0 ps, Offset Loss 1.3 Gohm/s)
REFLECT (SHORT, Delay 0 ps)
MATCH (@0 to 2GHz)
LINE1 (50-ohm transmission line, Delay 54.0 ps @ 2G to 7GHz)
LINE2 (50-ohm transmission line, Delay 13.0 ps @ 7G to 32GHz)
"
Please somebody help me to release my confuse
Have a nice day
MacGyFu
I make a set of coaxial cable TRL calibration kit. but I have a little confuse of defining the parameter of calibration kit.
I use Agilent E5071C network analyzer. The below example was from Agilent website. It tells me I need to define the offset loss of thru and line. But in my understanding, TRL theory should not need any known parameter of calibration kit, even the delay of line.
It is a little weird, If it is necessary, how do I define it? there are two different coaxial cable from NA to DUT, one is "expensive cable" which it was used to being reference plan of SOLT calibration. And the other one is TRL calibration kit cable. so I have two different cables if I need to measure offset loss. Do I measure cable1, cable2 or cable1+cable2?
"
Example of defining the TRL calibration kit
You need to enter the definition of the TRL calibration kit to perform TRL calibration. Follow these steps to define the following calibration kit given as an example.
THRU (Delay 0 ps, Offset Loss 1.3 Gohm/s)
REFLECT (SHORT, Delay 0 ps)
MATCH (@0 to 2GHz)
LINE1 (50-ohm transmission line, Delay 54.0 ps @ 2G to 7GHz)
LINE2 (50-ohm transmission line, Delay 13.0 ps @ 7G to 32GHz)
"
Please somebody help me to release my confuse
Have a nice day