Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

SOLT calibration procedure

Status
Not open for further replies.

elektr0

Full Member level 5
Full Member level 5
Joined
May 2, 2006
Messages
279
Helped
2
Reputation
4
Reaction score
3
Trophy points
1,298
Activity points
3,374
solt calibration procedure

Hallo,

for SOLT calibration, the standard 12term error model is used. Could somebody please explain why the transmission coefficient from the Network analyzer to the DeviceUnderTest is always assumed to be 1 (without loss of generality).

For better understanding I will download a picture of the 12 term error model.

Thank you.
 

Hallo,

Added after 1 minutes:

Hallo,

simple answer. E-TF covers both the transmission part of the error network 1 (which was set to 1) and the error network 2 (port2).
 