Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

simulate humidity sensor

Status
Not open for further replies.

Li_ereunpisan

Full Member level 5
Joined
Jul 14, 2009
Messages
260
Helped
32
Reputation
64
Reaction score
28
Trophy points
1,308
Location
Bandung, Indonesia
Activity points
2,761
I have a humidity sensor ..
how to simulate the temperature humidity as a substitute for the actual temperature of moisture
 

RCinFLA

Advanced Member level 2
Joined
Aug 31, 2010
Messages
606
Helped
187
Reputation
378
Reaction score
189
Trophy points
1,323
Activity points
5,788
Not sure what you are looking for. Humidity sensor is designed for relative humidity. If humidity sensor temp is below dew point I believe the humidity readout is invalid.
 

Li_ereunpisan

Full Member level 5
Joined
Jul 14, 2009
Messages
260
Helped
32
Reputation
64
Reaction score
28
Trophy points
1,308
Location
Bandung, Indonesia
Activity points
2,761
Not sure what you are looking for. Humidity sensor is designed for relative humidity. If humidity sensor temp is below dew point I believe the humidity readout is invalid.

I mean not because of an invalid ...
but I want to simulate to measure the temperature of moist, humid temperatures due to get very difficult
so essentially, how to change the temperature of humid ...
according to the theory ... mixing dry air and moist air flowing at a carefully regulated. but I am confused how it works and how to do it
maybe you can help

thanks
 

FvM

Super Moderator
Staff member
Joined
Jan 22, 2008
Messages
49,143
Helped
14,384
Reputation
29,032
Reaction score
13,139
Trophy points
1,393
Location
Bochum, Germany
Activity points
283,247
There are basically three methods for humidity calibration:
1. Adding defined water quantities to dry air or nitrogen and evaporate it.
2. Drying a wet atmosphere by cooling a surface to a defined temperature(dew point methode)
3. Using aqueous salt solutions with an atmosphere of known humidity above it, e.g. NaCl, KCl, LiCl as calibration standard. In contrast to saturated solutions, as listed in the below document, unsaturated solutions of known concentration have an almost temperature independant humidity atmosphere.

View attachment STANDARD SALT SOLUTIONS FOR HUMIDITY CALIBRATION.pdf
 