Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

[SOLVED] Question about DMMs and pulsed signals

Status
Not open for further replies.

d123

Advanced Member level 5
Advanced Member level 5
Joined
Jun 7, 2015
Messages
2,502
Helped
494
Reputation
992
Reaction score
525
Trophy points
1,393
Location
Spain
Activity points
27,128
Hi,

One DMM user manual says it refreshes the LCD every 2 to 3 seconds, but it is much faster than that regarding measurements. No information about sample rate.

Another DMM manual says the sample rate is twice a second, DC and AC from 50 to 300 Hz.

With a 100 Hz, 10% on:90% off duty cycle pulsed square wave signal, might the readings displayed on the DMMs be meaningless for DC currents and DC voltages?

e.g. I thought an 100mA 10% on, 90% off signal would read as 10mA, or a 1V signal as 100 mV, but am getting readings I do not understand such as 70mA.

What am I not understanding?

Possibly the problem is in part my not re-scaling/re-sizing the sense resistor and current clamp reference voltage adequately for a pulsed signal rather than for a pure DC signal for it to fit the goal (LED brighness, in this case)?

Thanks.
 

Carry for cents bazar

Member level 5
Member level 5
Joined
Aug 2, 2020
Messages
81
Helped
10
Reputation
20
Reaction score
16
Trophy points
18
Location
CZ
Activity points
417
Hi d123!

Please post circuits and part numbers when you ask questions. This makes it so much easier for us. I see you mean a digital multemeter.

I don't understand the question?

What do you mean by this:
With a 100 Hz, 10% on:90% off duty cycle pulsed square wave signal, might the readings displayed on the DMMs be meaningless for DC currents and DC voltages?

The manual means that it measures twice a second and you have to wait until the measurement is updated. If you have a 300Hz signal which changes 4 times per second to 20Hz, 50Hz, 200Hz and 300Hz, the multimeter will measure twice per second and not be able to display all the changes.
 
  • Like
Reactions: d123

    d123

    Points: 2
    Helpful Answer Positive Rating

ads-ee

Super Moderator
Staff member
Advanced Member level 7
Joined
Sep 10, 2013
Messages
7,942
Helped
1,822
Reputation
3,654
Reaction score
1,807
Trophy points
1,393
Location
USA
Activity points
60,185
Unless you have a good quality True RMS DMM then you shouldn't even attempt to measure a none sinusoidal signal with that meter. A 10% duty cycle square wave is not sinusoidal and won't give you meaningful results on a cheap meter.

Fluke does a nice job of why you need to use a True RMS meter for none sinusoidal measurements.
 
  • Like
Reactions: d123

    d123

    Points: 2
    Helpful Answer Positive Rating

KlausST

Super Moderator
Staff member
Advanced Member level 7
Joined
Apr 17, 2014
Messages
23,502
Helped
4,758
Reputation
9,537
Reaction score
5,172
Trophy points
1,393
Activity points
155,872
Hi,

There are various measurement methods currently used.
* average
* rectified average
* true RMS including DC
* true RMS excluding DC
None of them is right or wrong. Sometimes one needs the one, sometimes the other.

I thought an 100mA 10% on, 90% off signal would read as 10mA
100mA / 0mA 10% duty cycle, square signal will be*
* Average: 10mA = (10% × 100mA)
* rectified average: 10mA = (10% × 100mA)
* RMS including DC: 31.6mA = 100mA × sqrt(10%)
* RMS excluding DC: 30.0mA = sqrt(( 90mA × 90mA x 10%) + (10mA x 10mA x 90%))

Besides this there are many "non standard" measurement methods, mainly cheap DVMs where the measurement method gives correct results with some waveforms only, like pure sine without DC, square wave 50% duty

70mA may come from a "non standard" peak measurement method..

Klaus
 
  • Like
Reactions: d123

    d123

    Points: 2
    Helpful Answer Positive Rating

d123

Advanced Member level 5
Advanced Member level 5
Joined
Jun 7, 2015
Messages
2,502
Helped
494
Reputation
992
Reaction score
525
Trophy points
1,393
Location
Spain
Activity points
27,128
Hi,

Thanks, everyone.

One DMM for AC voltage is: 'Response: average responding, calibrated in rms, of a sine wave.'

No AC current measurement option, only DC current.

Other DMM for AC voltage and current is: 'AC conversion type: Average sensing rms indication. AC conversions are AC-coupled, true rms responding, calibrated to the sine wave input.'

I think from replies to the thread, neither DMM is adequate for this purpose without back-calculating with formulas Klaus provided.

Relevant respective pages attached as images, in case they are of interest.

Thanks a lot.


DSC_0003.JPG

DSC_0002.JPG
 

FvM

Super Moderator
Staff member
Advanced Member level 7
Joined
Jan 22, 2008
Messages
51,217
Helped
14,653
Reputation
29,584
Reaction score
13,797
Trophy points
1,393
Location
Bochum, Germany
Activity points
292,750
I understand the original question so that it's exclusvely about measurements in DC range. In so far the references to RMS capability are unrelated.

Correct measurement of pulsed DC current and voltage average value is a different problem. Unfortuntely it's rarely excactly specified in multimeter specifications.

In case of autoranging instruments, you should care that autoranging is disabled and a range according to the expected peak value is selected.
 
  • Like