Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Problem with using a single point load cell

Status
Not open for further replies.

arslankhalid

Newbie level 1
Joined
Aug 25, 2006
Messages
1
Helped
0
Reputation
0
Reaction score
0
Trophy points
1,281
Activity points
1,289
we want information about using a single point LOAD CELL
it is a load sensor. No matter how much load we apply the change in the reading remains negligable, the maker is TEDEA HUNTLEIGH model 1022. READING ALSO GOES NEGATIVE.
 

Re: LOAD CELL

More info is needed to help you. Please answer the following:

What excitation are you applying to it?

What type of circuitry are you measuring it with?

How is it wired to the measurement circuit?

How are you loading it?
 

Re: LOAD CELL

Bear in mind that the output of load cells is relatively small ..
In practice, most load cells output less than 10 mV/V, that is 10 mV of output per volt of excitation voltage ..
For example, with a 10 V excitation voltage, the maximum output signal is 100 mV ..
Therefore, load cell signal conditioners include amplifiers to boost the signal level and gain amplifiers with gains of 1000 and more are not unusual ..

Regards,
IanP
 

Re: LOAD CELL

88_1167006475.JPG
 