Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Oscilloscope Setting Based on Probe Specification

Status
Not open for further replies.

frodonet

Member level 2
Member level 2
Joined
Sep 2, 2011
Messages
44
Helped
20
Reputation
40
Reaction score
21
Trophy points
1,288
Visit site
Activity points
1,636
Hi All,

If my probe shows the following specs :

1) 10 :1
2) 1 Mohm for 6-24pF
3) 10 Mohm for 13pF

Even though my input technically is 50ohms from the source, does this mean i have to set my oscilloscope to 10x and input load to 1Mhom ?

By the way, what does (2) and (3) means?

Thanks
 

I essentially means that the oscilloscope input impedance must be according to 2). Probe compensation is possible for an input capacitance in the said range. The oscilloscope input must be 1 Mohm type, 10:1 setting is only a matter of scaling the display units correctly. 3) is the probe input impedance.
 