Oscilloscope probe voltage derating vs frequency

Status
Not open for further replies.

schmitt trigger

Advanced Member level 5
Joined
Apr 17, 2013
Messages
2,725
Helped
859
Reputation
1,722
Reaction score
850
Trophy points
1,393
Activity points
18,445
What is the technical reason which the scope probes are derated with frequency?

This is a generic probe from TPI, but have seen the same behavior on other brands, too
 

Attachments

  • IMG_E1542.jpg
    240.6 KB · Views: 309

Current handling of the compensation caps?
 

Frequency limit is unusually low. Many passive probes have full voltage rating up to 1 MHz.

I'd expect that voltage divider HV resistor as well as HV capacitor have high frequency derating.
 

yes, current at 600Vrms at 1GHz in the probe's capacitance would be high - also almost all caps have to be derated with freq - due to losses in the dielectric and loss of volt withstand with freq
 

From Tektronix application note Oscilloscope Probe Circuits by Joe Weber:

Reverse termination is referring to series resistor R2 in the below schematic, cable I²R is significant because passive probes are using a lossy cable with resistance wire inner conductor.

 
Makes sense now!

I was wondering how come a 600VDC probe would only withstand about 50V @ 10 Mhz, which is not that high in frequency.

Fortunately, logic circuits running that fast are all low voltage.

But it could be a problem in some modern SMPS powertrains.
 

The more typical derating curves of modern 10:1 and 100:1 probes should be sufficient for most SMPS applications.

 