Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

How do I calculate the Output of this potentiostat

newbie_hs

Full Member level 3
Full Member level 3
Joined
Mar 4, 2023
Messages
180
Helped
1
Reputation
2
Reaction score
2
Trophy points
18
Activity points
1,423
I am using this gas sensor in my circuit.The below circuit is obtained from vendors application note.

1)How do I calculate the voltage at the output of IC2
2)May I know the role of Rload

1744980384961.png
 
You can't calculate it, potentiostat compensates electrode polarisation voltage inside the sensor. Usually several dozen millivolts when sensor current is zero.
 
Last edited:
There isn't much to design. Application note suggests OP90/296 micropower bipolar OP, but any low offset, low noise FET OP can do as well.

Application note reports +600..+800 counter electrode potential for Chlorine sensor. Can be considered as first sensor health check, calibration with test gas is however required for a reliable function check.

Rload creates negative feedback for reference control loop and increases output stability.
 
From AN4348 by ST
The recommended load resistor of the sensor, specified in its datasheet, gives the value Rl .
Rl is a tradeoff between response time and noise. Indeed, any noise on the WE node is
amplified by the op-amp with a gain of 1 + (R T /Rl ).
Due to its capacitive behavior, the output signal converges with a time constant C s (R s +Rl ),
assuming that the sensor is a capacitor, C s , in series with a resistor, R s .
Also, Rl should not be too high or it may unbias the sensor (because of an Rl ×I sense voltage
voltage drop)
 
There isn't much to design. Application note suggests OP90/296 micropower bipolar OP, but any low offset, low noise FET OP can do as well.

Application note reports +600..+800 counter electrode potential for Chlorine sensor. Can be considered as first sensor health check, calibration with test gas is however required for a reliable function check.

Rload creates negative feedback for reference control loop and increases output stability.
I am confused about the circuit analysis of IC2 side.Suppose the voltage at reference electrode(left side of R1) is some X.How much will I get to counter electrode.
 
I am confused about the circuit analysis of IC2 side.Suppose the voltage at reference electrode(left side of R1) is some X.How much will I get to counter electrode.
This is an integrator. If there is a constant voltage at the input, output rises with constant speed
1745006046792.png
This circuit changes output voltage until there is zero at the input
 
1. Double RC is unusual, but doesn't change principal operation and DC levels. Don't know if has significant effect, may be reduces noise.

Don't remember to have ever used additional RC in potentiostat, not even with Alphasense.

2. Still zero.
 
1) why two RC sections.
Second section have ten times lower time constant. This attenuates frequencies far beyond crossover frequency. Looks like noise reduction.

Ic1 gives noisy signal because huge gain of IC1, I don't what part of this noise comes from integrator.
 
2. Still zero.
If the reference electrode is zero volt ,then why it is connected to counter electrode using integrator.
This section is not clear for me.
For an integrator When the input is zero the output also will be zero(assuming zero initial conditions).Please correct me if I am wrong
 