Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.
hi,
in INL &DNL testing you must have use histogram testing and so you must use as many sample as you have resolution in your ADC.
example, a linear ramp (in actual practice, a triangular waveform is used)which slightly exceeds both ends of the range of the ADC is a popular histogram test signal.
you can use a periodic ramp signal that is ramp proportional to your adc resolution and sampler frequency . for example if your INL need .05LSB (20 hit for each code )and you have 14-bit,100MSPS ADC ,so you have total NO of sample 20*2^14=327680 sample and sampling frequency 100 MSPS , this implies that the input ramp should make a full-scale transition in 3.3 ms.
I know the CDS test ,but in ADS or cadence , it takes so long time ,isn't it ? especially my ADC sample frequency is low 5Ms/s,Is there any simple way to do the test ?