Continue to Site

Welcome to EDAboard.com

Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Help me understand two configurations of ECG circuit

Status
Not open for further replies.

reyge

Member level 4
Joined
Jan 23, 2007
Messages
75
Helped
2
Reputation
4
Reaction score
1
Trophy points
1,288
Activity points
1,828
hello! Attached are two circuit configuration that are similar except that one uses a single supply voltage and some capacitor and resistor values. Can someone explain what the right leg drive do? How does the OP97 accomplish this? Will there be a difference between the two circuit configuration aside from the difference in the power supplies?

I could really use some help. Thanks!
 

ecgcircuit

Hello,

the "feedback" tries to zero common mode voltage, this way increasing the common suppression of the differential amplifier und surpressing interference, e. g. from AC power supply.

The second version is an non operating artefact without an appropriate input bias.
The designer should repeat his OP circuit class.

Regards,
Frank
 
ad620 artifact suppression

Hi Frank,

Thanks for the reply! The second version you are referring to is circuit 1.jpg, right? Can you elaborate on how it has no appropriate input bias? I attached the paper where I got the configuration. For the resistor values, do you think the authors just had a typographical error? R6 should be 1M and R5 should be 10k just like those in circuit 2.jpg


Also, the AD620 requires an input voltage range of (-Vs + 1.9) to (+Vs - 1.2). Does the right leg drive assure that the input from the right and left arm will be at this range? Thanks!
 
supply circuit ecg

probably it useful for you
 
ad620, circuits ecg

hi, very well, i was searchingin others sites, but here too much information of ecg
 

ad620 can not work ecg

oh i am really impressed awith this site!:D
 

ecg circuit protection

great bro it's really helpme so much.
but i've question to you
how that can connect to pc and view the result to pc...???
tanx before
 

resistor configuration for ecg

Thank
 

ecg ad620

i've design the ecg project with AD620 but i don't know it;s correct or not, can you tell me it's correct or not (in attachment).and what i must with the diodes?
tanx
 

right leg driven circuit

Diodes are for protection.

Circuit should work
 

basic ecg circuit single supply

so what series of diodes that i must use for that?
 

Re: supply circuit ecg

probably it useful for you

Could you please send me you link again
I couldn't see it here.
I want to know how it is possible to use AD620 in an ECG device just by one single supply.
Regards
 

Status
Not open for further replies.

Similar threads

Part and Inventory Search

Welcome to EDABoard.com

Sponsor

Back
Top