Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

ground clip on oscilloscope

Status
Not open for further replies.

gjo

Junior Member level 2
Joined
May 22, 2013
Messages
21
Helped
0
Reputation
0
Reaction score
0
Trophy points
1,281
Activity points
1,430
What purpose does ground clip on oscilloscope serve when odcilloscope is connected to 1 to 1 isolation transformer secondary?
 

You still need a reference current loop that is shorter
than chassis-wallplug-chassis if you want any decent
high frequency behavior.
 

oscilloscope grpund clip

When using an isolation transformer does it matter where you connect ground clip of oscilloscope? And why?
 

Re: oscilloscope grpund clip

You want to place the clip in the return path of the signal you are measuring, as close as possible to the tip so as to minimize extraneous ground currents from upsetting your measurement.
 

[Merged] still a little confused on ground clip pf oscilloscopes

Is it better to connect ground clip of oscilloscope near component i am measuring on secondary of isolation transformer or is it just as good to connect it at the voltage terminal of transformer secondary, woulnt that earth reference entire line?
 
Last edited:

It's better to put it near the component. If you put it at the transformer then any currents(from other circuits) in that path will upset your reading. Look at this way: you drain the sink in your house. If you look at the water going into your sewer line you'll only see what's coming out of your sink, but if you look at the sewer line back at the sewage treatment plant...
 
  • Like
Reactions: gjo

    gjo

    Points: 2
    Helpful Answer Positive Rating
Re: [Merged] still a little confused on ground clip pf oscilloscopes

Is it better to connect ground clip of oscilloscope near component i am measuring on secondary of isolation transformer or is it just as good to connect it at the voltage terminal of transformer secondary, woulnt that earth reference entire line?
Internally to your scope, the ground terminal of your leads is already connected to the safety ground of the power plug. Where that goes in your isolation transformer situation is not clear. If the scope was designed to be used only with an isolation transformer then it is possible that its ground terminal is also isolated from the power. But probably not. Before you go connecting the ground terminal anywhere, find out if that ground terminal is connected to any local ground. Just make sure that when you connect the ground lead to something in your circuit that you are instrumenting that you aren't shorting something out.

Usually when a line-powered scope is used to instrument a circuit, that circuit is either floating or ground-referenced to the same ground as the scope. If it is floating, then connecting the ground lead will force it to have the same ground reference as the scope. Even if the circuit is already ground-referenced, and you could theoretically omit the use of a ground clip, it is still a good idea to use it, connected to a ground point near the point being probed, as barry said.
 
  • Like
Reactions: gjo

    gjo

    Points: 2
    Helpful Answer Positive Rating