Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

How to measure 60dBm IP3 if the equipment can only test 50dBm IP3

Status
Not open for further replies.

cmosbjt

Full Member level 5
Joined
Apr 25, 2004
Messages
250
Helped
10
Reputation
20
Reaction score
2
Trophy points
1,298
Location
USA
Activity points
2,293
I am doing 2 tone IP3 test, using two signal source and spectrum analyzer. I can only get 50dBm IP3 from the test setup itself. But I know the part has 60dBm ip3 at least, how can i test IP3 higher than the test setup?
 

There are two problems. The two signal generators may be feeding into each other. The leveling loop will generate the sidebands. The spectrum analyzer may be degrading the measurement.
 

Also, if the part is an amplifier then you will have to look at the power supply regulation under two tone conditions to make sure you don't generate AM terms that will look like IMD.

You should be able to rule out the analyser as long as you use a good quality attenuator ahead of it and set the analyser's own attenuator carefully.

Also, can you find a signal combiner with higher isolation between the sig gen ports? Note that the match at the sum port of the combiner will usually dictate the actual level of isolation.

Does the device you are testing have a good match at the input? If not, then you may be getting less isolation between generators than you think. Sometimes you can add a small attenuator at the sum port to get the ideal compromise.

Also, have you checked to see if the IMD terms you are seeing are following the third order response when you change the drive levels? i.e. you may calculate a different IP3 for different drive levels and often devices that claim a high IP3 figure will quote the IP3 figure at a specific drive level. What (device output) test tone level are you currently using?
 