    How to measure differential signals using an Oscilloscope



    Differential signal measurement in practice

    To measure a differential signal, we have two options, one is using a differential probe and second is using a two channels oscilloscope. A differential probe is expensive but handles a better accuracy. Using two/four channels oscilloscope is the cheapest method which handles acceptable results.

    Just you need to connect oscilloscope’s Channel-1 to one of the differential lines/wires and Channel-2 to another one. Then go to the math function and enable CH1-CH2 which means a difference. Then adjust the oscilloscope to observe the signal. You use the run/stop or a single shot button to freeze the signal on the screen and examine it.

    I have examined the ADSL2+ signal on the telephone line using the above-mentioned differential measurement method.

    Reference: http://bit.ly/2Z7sDTt

    •   AltAdvertisement

        
       

  2. #2
    Advanced Member level 5
    Points: 30,422, Level: 42
    BigBoss's Avatar
    Join Date
    Nov 2001
    Location
    Turkey
    Posts
    4,459
    Helped
    1340 / 1340
    Points
    30,422
    Level
    42

    Re: How to measure differential signals using an Oscilloscope

    This method is a bit subjective and unstable.If the frequency is sufficiently high, phase shifts will be different by oscilloscope channels so the measurement may be a bit erroneous.
    You presume that the oscilloscope has %100 identical channels and Phase Shifts are linear but this is not true in reality..



    •   AltAdvertisement

        
       

  3. #3
    Member level 2
    Points: 2,466, Level: 11
    Achievements:
    7 years registered

    Join Date
    Sep 2007
    Posts
    43
    Helped
    1 / 1
    Points
    2,466
    Level
    11

    Re: How to measure differential signals using an Oscilloscope

    Quote Originally Posted by BigBoss View Post
    This method is a bit subjective and unstable.If the frequency is sufficiently high, phase shifts will be different by oscilloscope channels so the measurement may be a bit erroneous.
    You presume that the oscilloscope has %100 identical channels and Phase Shifts are linear but this is not true in reality..
    I am agree, as I mentioned in the article using the differential probe is the best, but not necessarily high voltage differential probes, because the main problem is to deal with high frequency pulses. This method at least gives a close picture of the final signal. The result on the screen is the result of the difference



    •   AltAdvertisement

        
       

  4. #4
    Advanced Member level 5
    Points: 14,971, Level: 29
    schmitt trigger's Avatar
    Join Date
    Apr 2013
    Posts
    2,488
    Helped
    805 / 805
    Points
    14,971
    Level
    29

    Re: How to measure differential signals using an Oscilloscope

    Also....one has to make sure that both probes are properly and equally compensated.

    A good way I have found is to compensate each probe individually (with the scope's CAL SIGNAL output).
    Then attach both probes simultaneously to the same CAL output and perform the A-B comparison, and slightly tweak for maximum cancellation.
    My batteries are recharged by "Helpful Post" ratings.
    If you feel that I've helped you, please indicate it as a Helpful Post



  5. #5
    Advanced Member level 5
    Points: 39,779, Level: 48

    Join Date
    Mar 2008
    Location
    USA
    Posts
    6,429
    Helped
    1878 / 1878
    Points
    39,779
    Level
    48

    Re: How to measure differential signals using an Oscilloscope

    Quote Originally Posted by schmitt trigger View Post
    Also....one has to make sure that both probes are properly and equally compensated.

    A good way I have found is to compensate each probe individually (with the scope's CAL SIGNAL output).
    Then attach both probes simultaneously to the same CAL output and perform the A-B comparison, and slightly tweak for maximum cancellation.
    I like to use the "signal of interest" (or one side of the pair)
    for this because the CAL signal on most 'scopes is way lower
    frequency / edge rate than most of what I look at.

    You would of course like to get identical model# probes to
    start with.



    •   AltAdvertisement

        
       

  6. #6
    Advanced Member level 2
    Points: 3,760, Level: 14
    Achievements:
    7 years registered

    Join Date
    May 2012
    Posts
    504
    Helped
    118 / 118
    Points
    3,760
    Level
    14

    Re: How to measure differential signals using an Oscilloscope

    On older oscilloscopes you invert one channel and sum the two channels. One problem with doing this is that the bandwidth is drastically reduced to around 1Mhz depending on the scope.



--[[ ]]--