    Probing with oscilloscope causes current drop

    We are testing a fabricated chip with expected output current of around 4mA which is then used to charge a button cell. The input is from a PV cell. However, when we attach an oscilloscope probe onto one of the pin outs of the chip (i.e. the ring oscillator and the clock), we detect a significant current drop at the output, from 4mA to around 1mA. The current from the PV cell also drops.

    what would be the cause of this current drop?

    We tried with a different oscilloscope, and still detected a drop in current, albeit of a different value.

    Attached is the block diagram of the DUT.

    implemented in 65nm CMOS

    note: the buffer is connected to the rectifier through offchip capacitors.

    Click image for larger version. 

Name:	buffer.jpg 
Views:	11 
Size:	131.3 KB 
ID:	153268

    •   AltAdvertisement

        
       

  2. #2
    Super Moderator
    Points: 74,899, Level: 66
    Achievements:
    7 years registered
    Awards:
    Most Frequent Poster 3rd Helpful Member

    Join Date
    Apr 2014
    Posts
    15,182
    Helped
    3460 / 3460
    Points
    74,899
    Level
    66

    Re: Probing with oscilloscope causes current drop

    Hi,

    We donīt see in the picture where you connect the scope.
    Also we donīt see what frequencies and what node impedances we have to expect.

    Thus itīs just guessing:
    * the capacitance of the probe may have influence on the frequency and/or the signal voltage...

    Klaus
    Please donīt contact me via PM, because there is no time to respond to them. No friend requests. Thank you.



    •   AltAdvertisement

        
       

  3. #3
    Advanced Member level 3
    Points: 5,198, Level: 17

    Join Date
    Feb 2014
    Posts
    719
    Helped
    245 / 245
    Points
    5,198
    Level
    17

    Re: Probing with oscilloscope causes current drop

    Yep, I'd assume its the parasitics of the probe, probably C.

    Put a 10pF cap on the same node and see if you get similar results.

    10X mode on the probe should be better than 1X.



    •   AltAdvertisement

        
       

  4. #4
    Full Member level 5
    Points: 3,019, Level: 12
    Achievements:
    7 years registered
    nitishn5's Avatar
    Join Date
    Mar 2011
    Location
    Bangalore, India, India
    Posts
    301
    Helped
    78 / 78
    Points
    3,019
    Level
    12

    Re: Probing with oscilloscope causes current drop

    As mentioned by others, the oscilloscope probe is loading your signal.

    I would suggest either using a high impedance probe and perhaps a high frequency probe that has low capacitance.
    But they would be costly.

    Another approach is to use a separate voltage buffer with high input impedance and low input capacitance only for probing. That way you can isolate the signals from the oscilloscope probe.
    -- Bilgotidia Fertiganed



--[[ ]]--