Problem with pyrometer
Hello everyone, at first sorry for my english I'm not good in it,
1 I have question about Infrared thermometer
2 I bought cheap pyrometer and it is I think uncalibrated or broken
3 Because their indications of themperature are not good
4 When I measure for example : my skin there is about 26 deegrees of Celsius
: boiling mug of water there is only 60 - 70 of C
: hot metal plate of iron there is only 80 - 90 nothing more (and hot iron has
: 150 - 200 C)
: hot metal plate on fire indicates only 120 - 130 C
5 And I think that it is ( in statum nascendi) broken or set wrong
6 I unscrewed this device and I found in it - interface ( I think) for calibration
7 And I would needed some hint from you guys how and if it is possible to manage this device with some simple external interface and some simple software?
8 I would like to get some already made schematic and link to something universal software from you.
9 I can make simple interface with passive elements like transistors or even with example: atmega 8 or other cheap IC's
10 Could you give me any help with my problem? ( my problem is not very important but I'm curious about it that it is possible to DO.
11 I know that I will be needed referencial temp meter to compare with my device.
12 I add photos of my device and internal interface.
13 Ending of my story again sorry for my language, but I think you had a lot of funny reading that,...
-
Advertisment
-
25th June 2018, 18:38 #2
- Join Date
- Jan 2008
- Location
- Bochum, Germany
- Posts
- 43,773
- Helped
- 13308 / 13308
- Points
- 251,373
- Level
- 100
Re: Problem with pyrometer
You can't use a pyrometer with bare metal surfaces, read he manual and learn about emissivity.
Try again with coated or blackened surfaces, if the reading is still wrong, the device may actually defective.
-
Advertisment
-
7th July 2018, 15:12 #3
- Join Date
- Jun 2018
- Posts
- 2
- Helped
- 0 / 0
- Points
- 79
- Level
- 1
Re: Problem with pyrometer
hmm, I just covered hot iron with dark paper and again tested this meter and I had on display only about 120 - 150 C and paper almost combusted, next step I have done - I took fireclay brick and put it on gas kitchen and warmed it up quite a lot and indications of temp were strange,. ok nevermind when I take some other meter I will compare then both,
-
Advertisment
-
7th July 2018, 17:01 #4
- Join Date
- Jan 2008
- Location
- Bochum, Germany
- Posts
- 43,773
- Helped
- 13308 / 13308
- Points
- 251,373
- Level
- 100
Re: Problem with pyrometer
Paper tag on metal surface should work, that's what I usually apply to measure bare heatsink temperatures.