Hello,
Please help to see of the probe here is matched to the TDS210 scope
I wish to probe SMPS circuits with the TDS210 oscilloscope...
TDS210 SCOPE:
**broken link removed**
The following probe is cheapest , so i pick it....
Picotech.com MI007 60MHz probe......
https://www.picotech.com/picoscope2200-accessories.html#MI007
So is this probe OK?
The scope has a bandwidth of 60MHz, so i presume it is a total waste of time to buy a probe with a bandwidth of any more than 60MHz?
Also, i am aware that when in 10X mode, the scope probe's capacitance must form a 10:1 divider with the input capacitance seen at the oscilloscope.
So, considering that the scope lead give about 90pF, and the scope input is 20pF, then thats 110pF at the scope input, so my divider capacitor in the scope probe must be 110/9 pF = 12pF
..the MI007 scope probe spec says it has 15pF, so therefore this is close enough to 12pF and will be ok?
Please help to see of the probe here is matched to the TDS210 scope
I wish to probe SMPS circuits with the TDS210 oscilloscope...
TDS210 SCOPE:
**broken link removed**
The following probe is cheapest , so i pick it....
Picotech.com MI007 60MHz probe......
https://www.picotech.com/picoscope2200-accessories.html#MI007
So is this probe OK?
The scope has a bandwidth of 60MHz, so i presume it is a total waste of time to buy a probe with a bandwidth of any more than 60MHz?
Also, i am aware that when in 10X mode, the scope probe's capacitance must form a 10:1 divider with the input capacitance seen at the oscilloscope.
So, considering that the scope lead give about 90pF, and the scope input is 20pF, then thats 110pF at the scope input, so my divider capacitor in the scope probe must be 110/9 pF = 12pF
..the MI007 scope probe spec says it has 15pF, so therefore this is close enough to 12pF and will be ok?