Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Matching oscilloscope probe to oscilloscope?

Status
Not open for further replies.

grizedale

Advanced Member level 3
Joined
Jun 13, 2011
Messages
838
Helped
17
Reputation
34
Reaction score
17
Trophy points
1,298
Activity points
8,804
Hello,

Please help to see of the probe here is matched to the TDS210 scope

I wish to probe SMPS circuits with the TDS210 oscilloscope...

TDS210 SCOPE:
**broken link removed**

The following probe is cheapest , so i pick it....

Picotech.com MI007 60MHz probe......
https://www.picotech.com/picoscope2200-accessories.html#MI007


So is this probe OK?

The scope has a bandwidth of 60MHz, so i presume it is a total waste of time to buy a probe with a bandwidth of any more than 60MHz?

Also, i am aware that when in 10X mode, the scope probe's capacitance must form a 10:1 divider with the input capacitance seen at the oscilloscope.

So, considering that the scope lead give about 90pF, and the scope input is 20pF, then thats 110pF at the scope input, so my divider capacitor in the scope probe must be 110/9 pF = 12pF

..the MI007 scope probe spec says it has 15pF, so therefore this is close enough to 12pF and will be ok?
 

The probe specification is incomplete, because it misses exact values for accepted oscilloscope input capacitance range. This may be due to the fact that the probes are primarly offered as acessory for Pico USB scopes. Most likely it works with all industry standard oscilloscopes, in case of doubt you can compare with PicoScope impedance specification.
 
also, the probe spec says it has 15pF of input capacitance, ...this must be wrong, because 1 metre of coaxial cable has at least 90pF of capacitance.?

- - - Updated - - -

also, the probe spec says it has 15pF of input capacitance, ...this must be wrong, because 1 metre of coaxial cable has at least 90pF of capacitance.?
 

It says 15 pF in 10:1 setting, that's in the usual range for passive probes.
 