    how to test earth connection

    I am using a XRF instrument and I want to be sure earth connection has sufficient quality since instrument has a lot of boards and any discharge can cause disaster. Could experts please advice how I can test my instrument earth connection quality with tools like Ohmeter or oscilloscope.

    Steve

    •   Alt1st August 2004, 06:32

      advertising

        
       

  2. #2
    Advanced Member level 1
    Points: 5,146, Level: 16
    sergio mariotti's Avatar
    Join Date
    Dec 2003
    Location
    Italy
    Posts
    468
    Helped
    67 / 67
    Points
    5,146
    Level
    16

    how to check earth connection

    I think you are speaking about the Heart connection on the mains supply line.
    If yes, witout a specific instrument, try this, but remember, the operation contain a risk.

    1) I assume Your mains is Line+Neutral.
    2) I assume the Neutral is "good enough" to have a potential near to 0 V and enough stable during the time.
    3) I assume that a some kind of GND contact on the socket is REALLY connected to the Heart, also if with a bad conection. THIS IS FUNDAMENTAL FOR SEFTY !
    4) Disable an eventual "differential" circuit breaker (or connect measurement before it).
    5) connect AC voltmeter between GND and Neutral (on the socket).
    6) ADVISE EVERY PEOPLE INSIDE YOUR BUILDING!!
    The following operation is RISK of FOLGORATION !

    7) connetc a power load (like a 100W lamp) between Line and GND. Lamp should turn on, read the ACVoltmeter (V1) and DISCONNECT SOON!

    8) Applying the Ohm law, the GND resistance is Rgnd=V1/(Power_lamp/V_mains)
    eg:
    V1= 6V
    Power_lamp=100W
    V_mains=220V
    Rgnd= 6/(100/220)=8.2 Ohm, that is good for most applications.



    •   Alt9th August 2004, 17:21

      advertising

        
       

  3. #3
    Advanced Member level 3
    Points: 4,930, Level: 16

    Join Date
    Jun 2002
    Location
    India
    Posts
    840
    Helped
    41 / 41
    Points
    4,930
    Level
    16

    how to test an earth connection

    Hi,
    check the earth(absolute) to neutral voltage if possible. This should be less than 10V. Also check the votages with respect to neutral. This should not be more than 10V from the pecification.
    Make sure the chassis is connected to the ground/neutral.